spider24SpectatorHi Mickey,
Kenneth Luk will be the first to test it at the end of October !
Olaf
(Edited by spider24 at 2:05 pm on Oct. 8, 2004)
Kenneth LukSpectatorHi Mickey,
Yes, I’ll be trying this diode out soon.
Will keep you posted.
Ken
mickey franklSpectatorHi Ken
looking forwards to hearing your opinion.
Interesting to see if this laser has variable pulse so proper perio treatment can be done.
spider24SpectatorMickey, if you like to test the unit for 2 weeks, i can arrange it. London is easy to reach and our CE-conformity covers GB ! Me or one of our guys could fly in the morning and leave in the evening or at the next day.
Olaf
Robert Gregg DDSSpectatorQUOTEQuote: from mickey frankl on 6:04 am on Oct. 9, 2004
Hi Kenlooking forwards to hearing your opinion.
Interesting to see if this laser has variable pulse so proper perio treatment can be done.Mickey,
I’d like to see a defined protocol and some safety studies before putting a 20,000 Hz diode into a pocket blind to treat perio.
I’m very concerned about any company that allows any doctor to “test” their laser w/o any training, or protocols or clinical science. Who has done research on this device and what are the parameters of operation? Who did the research and who has validated it?
Who is going to “test” the laser on? Patients? What happens if the “testing” goes wrong?
Ask the company for the science it has used to arrive at safe and effective parameters of care. Clinicians should not be put in the positiion of doing the company’s clinical research. That attitude has hurt–severely–a lot of pateints over the years. This is the company’s responsibility, not the willing dentists in the marketplace.
Your patients deserve it.
We are 15-20 years beyond the start of lasers in dentistry. There’ no excuse to skip clinical R&D.
Bob
spider24SpectatorI absolutly aggree Bob !
That´s the reason why me made and make studies at the Universities of Freiburg, Mailand and Vienna. The oldest studies are now running for 6 years. The oldest with 20 W / 10.000 Hz are from 1999. The new 30 W / 20.000 Hz unit is used for nearly one year at the RWTH Aachen, Univeritiy of Freiburg, Milano and Vienna.
If we offer “tests” with our devices the testing includes everytime a peronal laser safety instruction, technical and clinical instruction which takes about 4-8 hours depending on the skill of the operator. Not in a course with 10 other dentists.
We have a hotline which can be reached 12 hours a day.
Additionally, i have to say that we have the highest level of certification at elexxion (ISO 9001, EN 13485) and the highest level of technical and clinical certification to the laser unit which includes medical compliance testing.
You can´t work with wrong power settings because every possible application is named and has fixed power parameters which only can be changed +/- 30 %.
So what´s the problem
?
Ok, i stop offering tests of elexxion products in this forum because i don´t want to make marketing for our products.
But i think your wrong judging a company that philosophy and products you do not know.
Olaf
Robert Gregg DDSSpectatorOlaf,
No need to be defensive. I made no judgements. Nor am I am not trying to chase you off this board.
But you must understand that this board has some pretty sophisticated laser users who want more than just marketing claims from any company, or “drive by” drop offs of lasers for patient “testing”, or the testing of a device thought up in an engineer’s lab w/o proper clinical research.
I only asked questions regarding the understanding of the clinicians who would use any product w/o specific training and/or science.
That´s the reason why me made and make studies at the Universities of Freiburg, Mailand and Vienna. The oldest studies are now running for 6 years. The oldest with 20 W / 10.000 Hz are from 1999. The new 30 W / 20.000 Hz unit is used for nearly one year at the RWTH Aachen, Univeritiy of Freiburg, Milano and Vienna.
I think we would all like to see the results of your studies. 6 years is a long time to run a study, by the way. What parameters of clinical or research outcomes were you looking at? You are welcome to post any and all, as this group would like to see it.
If we offer “tests” with our devices the testing includes everytime a peronal laser safety instruction, technical and clinical instruction which takes about 4-8 hours depending on the skill of the operator. Not in a course with 10 other dentists.
OK, who does the training, a dentist or a sales rep?
Additionally, i have to say that we have the highest level of certification at elexxion (ISO 9001, EN 13485) and the highest level of technical and clinical certification to the laser unit which includes medical compliance testing.
You are talking about device manufacturing and Good Manufacturing Practices (“GMP” ) compliance. That’s great you have those certifications, but what does that have to do with the safety of patients when doctors are using your lasers w/o known clinical parameters?
You can´t work with wrong power settings because every possible application is named and has fixed power parameters which only can be changed +/- 30 %.
A clincian cannot apply the “right” clinical parameters if:
1. They don’t know what energy parameters they are using such that they can make intra-operative adjustments.
2. They do not know the energy at the fiber tip–unless it’s a “hot glass” effect.
3. The laser they are using has a variance of +/- 30% which is HUGE unless there are compensating conditions–like fiber “conditioning”.So what´s the problem
?
Nothing as long as you:
1. Present your science to back up your marketing claims.
2. Show data on safety and efficacy using your new parameters for the procedures you want dentists to test.
3. Train dentist thoroughly in the use of this new device parameters, before letting them experiment on live patients.Ok, i stop offering tests of elexxion products in this forum because i don´t want to make marketing for our products.
Perhaps a good idea in the absence of addressing the concerns I listed above.
Bob
lookin4tSpectatorBob..you sound like me now!
spider24SpectatorAll studies which are made until now are published in German language. But i post everything we get in english in the future.
One thing is very important: With the elexxion claros we are not talking about a protoype or something which has just left the lab !
There are ~ 400 units with 20W / 10.000 Hz and ~ 150 units with 30 W / 20.000 Hz which are succsessfull used in dental offices all over euope and asia for several years now.
“Testing” does NOT mean that any dentist should test if clinical parameters or treatment protocols are working. That was done at Universities years before. Due to the European law you have to show the evidence that our treatment protocols are right. But that documentation is not public.
“Testing” means to work a while with a well known unit to get an impression of the handling and the features.
We do not make any experiments with dentists or patients !
All energy parameters are clearly shown on the display. And an internal calibration system measures the output power continueisly. When there are differences more than 10 % between the real output power and the displayed power the laser will switch off with an error message. Intra-operative adjustment is possible with +/- 30 % from the original value to compensate power changes due to fiber conditions.
Olaf
Robert Gregg DDSSpectatorOlaf,
Thank for the reply. Now we are getting somewhere. Answers as a laser clinician I can begin to understand.
All studies which are made until now are published in German language. But i post everything we get in english in the future.
You can provide the papers to my partner, Dr. Del McCarthy who can read German. Or you could write a summary of the critical findings.
One thing is very important: With the elexxion claros we are not talking about a protoype or something which has just left the lab !
Thanks for clarifying.
There are ~ 400 units with 20W / 10.000 Hz and ~ 150 units with 30 W / 20.000 Hz which are succsessfull used in dental offices all over euope and asia for several years now.
Excellent! Then you ought to be able to collect some clinical performance data from these clincians and publish a retrospective study on clinical uses and adverse outcomes (if any).
“Testing” does NOT mean that any dentist should test if clinical parameters or treatment protocols are working.
That is very reassuring to know.
That was done at Universities years before. Due to the European law you have to show the evidence that our treatment protocols are right. But that documentation is not public.
Why isn’t the documentation public? If it is university research, then it is science. If it isn’t science, then what is it? If you are not willing to make it public, then I start to get suspicious about the marketing and clinical claims with “secret” documentation.
“Testing” means to work a while with a well known unit to get an impression of the handling and the features.
How much training on an unfamiliar laser will a dentist get before they “work a while with a well know unit” except it isn’t well know to them, nor is the safety and efficacy documentation made public. Or is it?
We do not make any experiments with dentists or patients !
That’s very good to know.
All energy parameters are clearly shown on the display. And an internal calibration system measures the output power continueisly. When there are differences more than 10 % between the real output power and the displayed power the laser will switch off with an error message. Intra-operative adjustment is possible with +/- 30 % from the original value to compensate power changes due to fiber conditions.
Sounds like a nice feedback mechanism. The first thing I would do is put a power meter up to your fiber and see if it is within the specs you cited. That’s how I would first “test” your laser. I would do a few more tests, but not on humans, before I would work with a device whose parameters I was not familiar with.
Thanks for clarifying your company’s device.
Bob
mkatzSpectator6 months or so have passed since the last post – Has Elexxion yet provided copies of their published studies?
Robert Gregg DDSSpectatorQUOTEQuote: from mkatz on 7:47 pm on May 28, 2005
6 months or so have passed since the last post – Has Elexxion yet provided copies of their published studies?Nope……..
spider24SpectatorSorry, the results are not published until now – but it will be very intersesting. I´ll keep you informed.
olaf
mkatzSpectatorSpider24: in October, 2004, you wrote ” All studies which are made until now are published in German language.”. Why not mail/e-mail/fax copies of the German language published articles to Dr. Gregg so that hios partner, Dr. McCarthy can read them? My understanding is that Dr. McCarty is bilingual in German and English.
spider24SpectatorComparative Study on CW-Mode versus Pulsed Mode in
GaAlAs-Diode-LasersDr. med. Dr. med. dent. Claus Neckel*, D-Bad Neustadt / Germany
ABSTRACT
In the last six years, GaAlAS-diode-lasers have become more and more popular. Due to their small size, their good electro-optical coupling and delicate glass fibers, this type of laser fits into most dental offices perfectly. The first diode-lasers and still most of the devices on the market, work in a continuous wave mode or in a gated mode up to 50 Hz. When using this setting, high temperatures in the tissue are inevitable.
In this randomized study we tried to evaluate the difference in clinical cutting efficiency, post operative outcome and the histological findings of the excisional biopsies using a new diode-laser called ora-laser jet 20.
We preformed 40 biopsies:
Group I: in cw-mode
Group II: in pulsed mode 10.000 Hz gated (1:1)
Group III: in pulsed mode 10.000 Hz gated (1:3)
Group IV: in pulsed mode 10.000 Hz gated (1:10)
The macroscopic results and postoperative outcome of the patients in group IV were best.
Scarring and postoperative pain was more intense in the cw-group and high-pulsed group with an on/off-relation of 1:1. the histological evaluation showed more thermal damage the longer the laser pulse lasted.Keywords: laser, diode, cw-mode, pulsed mode, gated on/off-relation
1. INTRODUCTION:
A surgical diode-laser in dentistry was first introduced in 1994 by ORALIA, Germany1. Since that time, many companies have seen its potential and brought many new products on the market due to the main advantages of this lasertype as: small size, good electro-optical coupling and the use of delicate glass tips or fibers. On the other hand, there are still some technical pitfalls that have to be overcome.
Diode-lasers deliver their energy in a continuous wave mode or a gated pulsed mode. Thermal interaction with the lasered tissue is largely dependent on the wave length, the target tissue, the output power at the fiber tip and the application time.
Heat dissipation or diffusion from the site is a function of the thermal conductivity of the tissue and laser time. Collateral damage in the surrounding structures can ultimately be observed after laser application. Minimal thermal side effects are the aim of laser application to differentiate from sheer thermal cutting using electro-surgery.Due to the technical problems, diode-lasers run either on a continuous wave mode or gated2, mostly in an 1:1 on/off-fashion. Using this kind of application parameters, we experience a high thermal interaction and a continuous temperature rise in the target tissue. Low output powers are necessary to keep charring and thermal damage minimal. Cutting efficiency is reduced and intraoperative anesthesia is mandatory.
In 1999, a high-pulsed diode-laser (ora-laser voxx 08) was introduced which could achieve a gated pulse frequency of up to 10.000 Hz. With a laser diode rendering 8 Watt output power at the fiber tip, the cutting efficiency3 could be accelerated, nevertheless not without giving users problems in finding the right parameters for a sufficient application of the laser.
Aim of this study was to use fixed parameters of a new diode-laser for excisional biopsies of the oral mucose and to evaluate the clinical, individual and histological outcome of the patients and the specimens.
2. MATERIAL AND METHOD
In the study, we included 28 female and 12 male patients in an age range between 34 an 71 years. All 40 patients were randomly selected. The excisional biopsies were preformed under local anesthesia, using 3 % Aricain with 1:200.000 Adrenaline (Ultracain&Mac226; DS). Aim of the therapy was to obtain a histological diagnosis and to remove the entire tumor.
2.1. Clinical diagnosis of the biopsies:
26 fibromas
14 epulides fissurataThe dimensions of the tumors were in a range of 5 cm to 8 mm and were primarily situated in the cheek or the vestibule.
After excision, no sutures were used and no wound dressing was applied. If necessary, the denture of the patients with epulides fissurata was relined.2.2. Our laser:
The laser we used in the study was the ora-laser jet 20 of the company ORALIA GmbH, Konstanz/Germany. It is a GaAlAs-diode-laser emitting 810 nm. Maximum output power was 25 Watt. Continuous wave mode and pulsation of up to 10.000 Hz gating with an on/off-ratio between 1:1 and 1:10.
2.3. Clinical parameters in the study:
Group I:
400 µm glass fiber, continuous wave mode, 1.5 Watt output power at the fiber tip.Group II:
400 µm glass fiber, pulsed mode 10.000 Hz, gated 1:1 on/off-relation; a peak output power of 5 Watt at the fiber tip.Group III:
400 µm glass fiber, pulsed mode 10.000 Hz, gated 1:3 on/off-relation; a peak output power of 20 Watt at the fiber tip.Group IV:
400 µm glass fiber, pulsed mode 10.000 Hz, gated 1:10 on/off-relation; a peak output power of 25 Watt at the fiber tip.Cutting efficiency and intraoperative appearance was monitored. All patients were followed-up after 3 days, 1 week and 4 weeks. We evaluated all patients for their individual postoperative pain, clinical aspects of wound healing and the histology of the specimens with respect to thermal damage to the tissue.
3. RESULTS
3.1. Operative findings:
Cutting efficiency:
The comparison of cutting speed showed that the continuous wave mode with 1.5 Watt output power at the tip had the lowest praparation rate.
It was followed by the biopsies of group IV with the on/off-ratio of 1:10. The difference between both groups was significant, whereas the speeding up between group IV and III (1:3 on/off-ralation) was only subjective. The significantly fastest group were the biopsies of group II.Clinical aspect of the donor site:
The parameters we looked at doing the excisional biopsies were charring and blanching of the tissue. Group I (cw-mode) showede some charring of the epithelial lining. The tissue had a very dry aspect and about 0.5 mm blanching around the excision. Group II looked very similar to group I. Charring was also significantly present. Both group III and IV presented different features. Charring was significantly less visible and the surface of the wound had a raw less dry aspect.Coagulation:
All groups showed sufficient coagulation in excision region. No bleeding occurred during the procedure or the wound healing phase even in patients with bleeding disorders (N: 6; 5 Acetylsalicylic Acid, 1 Coumarintherapy). The wound surface was totally dry in group I and II and showed a raw aspect with bleeding spots in group III and IV. These bleeding spots did not require any treatment and showed no bleeding tendency even after vigorous rinsing.3.2. Postoperative findings:
Individual patient outcome:
All patients reported of no or minimal postoperative discomfort. In those cases where the excisions were in direct contact to the margin of the dentures, the patients observed discomfort during masticatory function. No significant difference between the groups could be evaluated.Clinical postoperative follow-up:
At the three day postoperative inspection in group I, we found a slight redness of the direct surroundings of the excision. The wound had a grayish-white layer of fibrin on the surface. Wound sensitivity was minimal. One week postoperative, we found closing of the wound from the margins. Even small wounds of 8 mm in diameter had not yet filled in. Redness was not present. After four weeks, all wounds had closed in. In most cases, it was very difficult to distinguish the excision site. Scarring was minimal.Group II showed significant redness of the surrounding tissue after three days. Wound sensitivity was minimal even with the showing redness. The fibrin layer on the surface of the wound showed typical structure. At one week, in three cases redness of the surrounding tissue still without sensitivity was present. No wounds had yet closed in. The wound healing seemed to be delayed in comparison to group I. After four weeks, all wounds had healed. The identification of the excision site was easier as little scarring was detectable. The scarring was a visible mark without functional changes as induration or tissue augmentation.
Group III displayed minimal redness around the excision after three days. Fibrin layer and wound sensitivity were normal. After one week, the wounds were closing in. No wound, even small ones, had totally reepithelialized. After four weeks, all wounds had healed. Almost no scarring was distinguishable.
Group IV presented only in two cases redness in the periphery of the wound after three days. Sensitivity was minimal. All wounds had a fibrin layer. At the one-week follow-up, we found good wound healing with a significantly faster epithelisation rate. A small wound had almost closed in. After four weeks, the excision sites could not be distinguished. Scarring was not visible.
3.3. Histological findings:
The histological evaluation of the biopsies confirmed in most cases our clinical diagnosis. Only in one case, the diagnosis fibroma hat to be specified as ossifying fibroma. In the specimen you could find osteoclastes and osteoblasts with mineralized and non-mineralized osseoid. The tumor had no hint for malignancy and was removed totally.
In group I with continuous wave mode, the histology showed a thermal damage zone of up to 125 µm. The range found lay between 80 and 125 µm. The target tissue for thermal damage is connective fibrous tissue. Very little muscle cells are effected.
Dehydration of all damaged cells is prominent. Signs of coagulation and protein denaturation are predominant. Some superficial charring adjacent to deeper thermally deteriorated zone could be found.Group II exhibited similar results only that the thermal effects were significantly more drastical. The thermal alteration showed a depth of 250 to 800 µm. Muscle cells and fibrous tissue showed dehydration and cellular disintegration, Charring was an often found site with very deep thermal damage.
In group III the histological changes exhibited a thermally damaged zone of 45 to 65 µm. Connective tissue again was the main target. Dehydration of the damaged cells was prominent. Charring was found very seldom and the thermal destruction zone was more constant in depth.
Group IV displayed a maximum of thermal damage of 25 µm. The depth ranged between 15 and 25 µm. Charring was not visible. The thermal damage zone was very consistent in depth.
All specimen could be evaluated by the pathologist. Total marginal evaluation was only possible in the specimen of groups III and IV and of group I in 50 % of all cases.
4. DISCUSSION
Photothermal interaction with tissue is the basis of surgical lasers. In this process, radiant light is absorbed by the tissue and becomes transformed to heat energy changing tissue structures4. The amount of laser light that is absorbed is dependent on a number of factors such as: wave length of the laser radiation, power output at the laser tip or focus, pulse or non-pulse , if applicable: pulse character and the optical properties and composition of the target tissue5. Every single parameter of the laser beam can significantly alter the thermal interaction of the laser radiation in the tissue. The wave length by large determines the heat generation in tissue as the absorption of the radiation in a specific tissue is dependent on the emitted photon character. The CO2-laser-beam is highly absorbed by intracellular water with soft tissue. Because of this, 90 % of its energy is absorbed within the first 100 µm of penetration of the tissue surface3. The high absorption of the laser energy leads to rapid vaporization of water and pyrolysis of organic material in the target tissue. High superficial carbonization and charring can be observed even when using low energy densities. Diode-lasers emitting 810 nm are well absorbed in dark pigment as hemoglobin, melanin pigment and other dark tissue components. The absorption rate in water is by far lower than in a CO2-laser6,7.
Laser light is primarily absorbed by the tissue inducing a molecular excitation8. Important is that the target molecules have a similar energy level structure as the laser wave length. Later the thermal energy is dissipated to the surrounding tissue structures by thermal diffusion and conduction. Heat dissipation is dependent on the thermal conductivity of the target tissue and will determine the extent of collateral damage in the adjacent tissue. This process which is highly influenced by the vascularity of the tissue, is characterized by the thermal relaxation time of a specific tissue. The degree of collateral damage is dependent on the target tissue properties such as tissue composition, water content, vascularity, volume of radiated tissue and tissue surface9. Furthermore, radiated energy and length of radiation are important factors in this process. Thermal relaxation time is the time that is required for a specific tissue to cool down from an accumulated temperature to 37 % of the maximum value10.
Using a continuous wave mode for surgery thermal dissipation is minimal other than by blood flow. Constant motion of the fiber tip is therefore mandatory to avoid overheating of the tissue11. Although lower power output is recommended, tissue dehydration and charring is observed. Using a pulsed approach, in a gated 1:1 on/off-mode, tissue has time to dissipate heat and to keep the temperature from rising continuously. In our clinical setting using a 10.000 Hz pulsation, we encountered more clinical, individual and histological damage than the cw-mode group showed. This is due to the fact that the time of the off-phase was too short for a sufficient heat conduction, leading to a continuous increase of temperature. This also explains the clinical signs of mild inflammation in the postoperative phase and the scarring of the healed tissue. By lengthening the off-phase, heat dissipation is much more sufficient and thermal damage can be controlled much better. This can only be achieved at the expense of cutting speed. Therefore the pulsed mode with an on/off-relation of 1:1 had the best cutting momentum, an on/off-ratio of 1:3 was similar and the on/off-ratio of 1:10 showed a slightly slower speed. High output power at the fiber tip helps cutting efficiency and with clinical parameters of 20 or 25 Watt the cutting is comparable to a CO2-laser.
Important when comparing diode lasers is the fact that we defined the output power as power at the fiber tip and not as output power of the laser diode. Due to the ovoid outlet, the optical coupling is very difficult and often we find a significant loss of energy between laser diode and fiber tip12.
In the histological specimen, we found equivalent results to the clinical findings. Using a very high pulsation rate and short relaxation time, a thermal effect in the tissue is prominent. Charring can be found readily. This carbonized tissue is an example of laser induced modification of taget tissue. By burning the superficial tissue, a new extremely high absorbing agent is created that can create excessive high temperatures of up to 3200° C. This temperature is thereafter dissipated and conducted into the surrounding tissue damaging its structures irreversibly. All specimen showing charring simultaneously exhibited a deeper concentric thermal damage zone. By using a very high power output and longer off-period in the gated pulsed mode, thermal interactions in the tissue could be controlled. Charring was minimal and therefore unintended side effects of excessive heating as could be found in the gated one-to-one on/off-relation.
By controlling the thermal side effects of laser application, we found that this treatment stand out again electro-surgery. In this therapy, thermal destruction is by far more prominent. A further advantage of both treatments is their coagulating ability. Even in group IV with minimal thermal destruction of the target tissue, the coagulation was absolutely sufficient to achieve a stable wound over the period of healing. This therapy can also be used for patients with bleeding disorders3.
5. CONCLUSION
After introduction of high-pulsed diode-lasers, it has become very difficult to distinguish the right clinical parameters for the surgical use of these machines. For excisional biopsies with the ora-laser jet 20, we recommend using a continuous wave mode with low output power and low cutting efficiency or to excise with high output power, high pulsation rate and at least an on/off-ratio of 1:3, better 1:10. Clinically and histologically, 25 Watt peak output power at the tip, 10.000 Hz and an on/off-ratio of 1:10 gave the best results. The difference in cutting momentum was not significant. 10.000 Hz combined with an on/off-ratio of 1:1 brought the largest thermal destruction zone and should not be used. All clinical parameters apply to the ora-laser jet 20 and do differ from other diode-lasers on the market. They should be adapted adequately.
-