    Great! I am having enough loss of sleep just trying to make the decision between DeLight and Waterlase! Thanks Kenneth! smile.gif
    I would love to hear more on this from the pros here also.

    #7195 Reply

    ericbornstein
    Spectator

    Kenneth:

    Very excellent post.

    First, Niemz clearly brings out, that with Erbium energy, until you get below 1 microsecond in pulse-width, you are dealing with thermal-mechanical ablation of dental tissues. So as far as the ablation mechanism is concerned, 80 microseconds is thermal, as is 200 as is 400.

    The area to look out for (in my opinion) is the Hz. At smaller pulse widths, and higher Hz (up tp 50Hz as you said in your post) you need to worry about pulse-stacking and heat build up.

    My opinion would be, that as long as you keep enough water on the ablation site during irradiation (heat sink) you may be OK.

    I would like to see a reliable study that would say something like:

    “80 microsecond pulse-widths at 40Hz cuts faster with no apparent thermal consequence than 200 microsecond pulse-widths at 20Hz”

    If Fotona could produce such data, in the face of the available literature stating that excess heat will inhibit ablation, I would like to see it.

    With such data, I would take a serious look at thier claims. Without it, I would be very careful with these energies, at these pulse widths, pumping up the Hz to 40 or 50Hz, especially in Bone.

    I am interested in other opinions please.

    Eric Bornstein DMD

    #7233 Reply

    Kenneth Luk
    Spectator

    Hi Eric,

    Great to hear from you! I was so much looking forward to your post on this thread.

    I’ve spent very long hrs last night going through your discussions with Ray on HK. Your knowledge on the topic enlighten us all. Thank you for sharing your views with us. I don’t want to dwell into that but like to say I use HK as short for Hong Kong ( :biggrin:). I’ll have to be careful in my future posts to avoid using it.

    BACK TO FOTONA
    I certainly would ask for more details on the parameters recommended for different procedures. Will post more info once I get them.

    Correct me if I’m wrong:

    The shortest available is 80us which is more than half as long as its closest rival. This allows the user to get a very short pulse of high peak power which gives very efficient cutting of enamel without requiring a large amount of energy per pulse to gain effective cutting.

    I thought he meant that by shortening the pulse width, less energy is required.

    Being able to vary the pulse width as well as Hz and W , won’t we be able to control thermal relaxation more? As Bob Gregg mentioned on the difference in free running pulse and gated pulse between Nd:Yag and diode. Obviously Er penetration is so much less than Nd:Yag and diode and we’re talking about hard tissue here instead of soft tissue. Does this still apply?

    Regards,

    Ken from HK (Hong Kong) 🙂

    #7198 Reply

    ericbornstein
    Spectator

    Ken:

    Thank you for the kind words.

    You are correct in your assumptions, as long as you understand the energies (and I believe yo do).

    If however a dentist buys this FOTONA Er:YAG, and say does the following, by only listening to a salesman:

    1) Turns up the watts to high
    2) Turns the pulse-width down to 80 microseconds
    3) Turns up the Hz to 50 (in effect double the energy from 25Hz)

    Then you have maximal power output (probably faster cut), with a phenomenon known as pulse-stacking, that may make the extra 60 or 70 micrseconds of thermal relaxation time in the pulse-width irrelevent if the heat goes up significantly at the ablation site, for a variety of reasons already discussed on the DT board.

    I would love to see some hard data describing the lower pulse width with higher Hz and power output, to see exactly how and if the heat is a problem.

    Nice post

    Eric Bornstein DMD

    #7236 Reply

    Kenneth Luk
    Spectator

    Hi Paul,

    Some comments from other competitors.

    In comparison to the Waterlase and the DElight my assessment would be as follows:

    – When cutting or removing dentine and caries, our unit is the same as the DElight. And both units are superior to the Waterlase for these applications.

    – For cutting enamel the Waterlase is superior to the DElight and our unit is considerably faster than both of them.

    – The Waterlase is the best of the three units for soft tissue, but it is a compromise, it is ok for soft tissue but not as good as your diode or the KTP. I know that Biolase have lots of video footage showing haemostatic surgery but as soon as the tissue is vascular or inflamed it will bleed. As you would know, with the diode this is not the case. And the KTP is even more haemostatic again.

    Ken

    (Edited by Kenneth Luk at 1:25 pm on Sep. 23, 2003)

    #7212 Reply

    Albodmd
    Spectator

    Ken,
    Was there any research to support those statements about the lasers or was that opinion? I would love to see the literature that supports those statements.
    Al B

    #7226 Reply

    Kenneth Luk
    Spectator

    Hi Al,
    That was just comments made by the rep.
    I’ve already asked for some papers and data about VSP and parameters for various hard tissue ablation.
    Will post more once I get them.
    Ken

    #7224 Reply

    Kenneth Luk
    Spectator

    Hi Eric,

    I’ve just received the document on VSP. Not had a chance to read it but want to post it for your comments first.

    The protocals for using VSP is as follows:

    VSP 100us ; short pulse 300us ; long pulse 600us ; very long pulse 1000us

    Enamel – Very Short Pulse 400mj and 10hz
    Dentine – Short Pulse 300mj and 10hz
    Caries – Short Pulse 120 – 220mj and 10 -15hz
    Bone – Very Short Pulse 200 – 400mj and 10hz
    Soft Tissue – Long Pulse 100-200mj and 15 – 20hz

    The enamel cutting setting is Very Short Pulse at 4 watts. This in only 40% of the lasers potential power output (which is 10 watts) but we don’t use more than four watts because it as fast as you would want or need to use the laser for cavity preparation instead of a drill.
    _____________________________________________
    VSP Doc.

    In vitro study of intrapulpar temperature change with
    short and very short pulse Er:YAG laser preparation of class V cavities

    Brugnera Junior, A.1; Zanin, F.1; Marchesan, M.A.2; Silva, R.S.2; Pécora, J.D.2.
    1Laser Center – Camilo Castelo Branco Uníversity; APCD and Federal University of Rio de
    Janeiro -UFRJ – Brasil.
    2Dental School of Ribeirão Preto, University of São Paulo – Brasil.

    High-density lasers such as CO2 and Nd:YAG generate heat in tissues. This is a concern for researchers because this heat which is generated during cavity preparation can cause irreversible changes in the dental pulp as well as periodontal necrosis (Zack and Cohen, 1965).
    Er:YAG laser, because of its interaction with water, causes dental tissue ablation without causing injuries such as fusion, cracks and carbonization (Keller and Hibst, 1995; Watanabe, 1998). Thus, the Er:YAG laser has been indicated for the removal of caries, cavity preparation and preparation and disinfection of root canals.
    Various studies have recommended parameters for the use of Er:YAG laser for the treatment of caries of the enamel and dentin without studying the different increases in temperature caused by the laser in the individual teeth, which because they have different anatomic characteristics, present different thermal diffusion responses.
    The authors studied the intrapulpar temperature change in 10 incisors, 10 canines, 10 pre-molars and 10 molars during Class V cavity preparation with short pulse (250 ms/pulse) and very short pulse (80-120 ms/pulse) Er:YAG laser, using the following parameters: 10 Hz frequency, 500 mJ potency, 6 s, 10 mm distance, 25 m1/min water flow, at 230C and 65% humidity.

    The greatest increase in temperature was found in the incisors and the least increase in the molars at both pulse modes. The very short pulse mode caused less of an increase in temperature in the pulp chamber in all teeth than the short pulse mode.

    _____________________________________________

    In Vitro Study of Cutting Efficacy of Very Short Pulse (VSP)
    and Short Pulse (SP) Er:YAG laser cavity preparation
    Brugnera Junior, A.1; Marchesan, M.A.2; Silva, R.S.2; Zanin, F.1; Guerisoli, D.M.Z.2,
    Pécora, J.D .2.
    1 Federal University of Rio de Janeiro —UFRJ – Brasil, and
    Laser Center – Camilo Castelo Branco University.
    2 Dental School of Ribeirão Preto, University of São Paulo – Brasil.
    A new pulse for Er:YAG laser (very short pulse, VSP: 80-120 ms/pulse) has been developed for faster cutting with less thermal damage. This study evaluated, in vitro, the cutting efficacy of VSP and SP (250 ms/pulse) Er:YAG laser cavity preparation. A total of 20 hemisections of molar crowns were placed in an incubator at 37ºC and weighed on a precision scale (MLW, Germany). After the 3rd measurement the weight stabilized and this weight was used as the initial sample value. A cavity was made in each sample using Er:YAG laser (Fidelis model 32OA, Fotona Medical Lasers, Slovenia) at both VSP and SP. The parameters used were: 10 Hz frequency, 500 mJ potency, 6 s, 10 mm distance, 25 ml/min water flow, at 23ºC and 65% humidity. The teeth were then dried with absorbent paper and placed in the incubator at 37ºC. The teeth were weighed and after stabilization of weight the difference in mass was calculated. The results showed that VSP Er:YAG laser was more efficient for cavity preparation than SPEr:YAG laser.

    _____________________________________________

    Er:YAG Laser Cavity Preparation with Variable Pulsewidth:
    A Microleakage Study.
    R.S. Navarrol; S.C.Gouw-Soares1; P. Haypek1; D.M. Zezell2; M.N. Youssef1, C.P. Eduardo1.
    1Faculdade de Odontologia-FOUSP; 2Centro de Lasers e Aplicações-IPEN/CNEN-SP,
    São Paulo- Brazil
    The Er:YAG laser (2.94,mm) is highly absorbed in both water and hydroxyapatite has great applicability on cavity preparation. The purpose of this study was to evaluated in vitro the microleakage in composite restorations performed by Er:YAG laser. Identical size class V cavities were prepared at the buccal surfaces in twenty-five bovine teeth by Er:YAG laser (Fidelis 320A, Fotona, Slovenija) with irradiation parameters 350mJ/ 4Hz or 400mJ/ 2Hz and pulse width: group 1: VSP- Very Short Pulse; group 2: SP- Short Pulse; group 3: LP- Long Pulse; group 4: VLP- Very Long Pulse and group 5: by high-speed drill (control group). All cavities were etched with 35% phosphoric acid for 15sec, washed with water, treated with adhesive system (Single Bond-3M), filled with composite resin (Z-250-3M). The specimens were stored at 37ºC in distillated water and polished 24 hours later with Sof-Lex discs (3M) and thermally stressed (700 cycles/ 5(-55ºC). The teeth were impermeabilized, immersed in 50% silver nitrate solution for 8 hours, embedded in acrylic resin and sectioned longitudinally with Isomet saw and exposed to Photoflood light (G.E.) for 10 min to reveal the stain. The specimens leakage were observed under stereomicroscope, SEM, and were double blind evaluated with scores (0-3), by 3 different examiners. The results were analyzed with Kruskal-Wallis test (p(0.05), showing no significant differences between tested groups. The morphology of cavities performed by laser showed irregular cavosurface enamel and internal walls, with more conservative pattern than conventional cavities. The laser irradiation parameters and pulse width used in this study for cavity preparation with Er:YAG have no influence on microleakage.
    Financial Support: CNPq; CAPES

    _____________________________________________

    Regards,

    Ken

    (Edited by Kenneth Luk at 7:35 am on Sep. 24, 2003)

    #7199 Reply

    ericbornstein
    Spectator

    Ken;

    Thank you very much for posting the study results.

    I have to tell you, that I am impressed. From a thermodynamics perspective and a thermal-mechanical ablation perspective, the results make sense.

    If you put higher power into a shorter pulse, with moderate Hz, (20 ish Hz), and keep the area cool with a water spray on the ablation site, you should get better results from both a “lower thermal deposition” perspective and a “better “explosive” ablation” perspective. Absolutely follows the logic of what we are trying to accomplish.

    I am definately going to take a look at the machine at the next show. Very Nice Post.

    Please keep it up.

    Eric Bornstein DMD

    #7192 Reply

    drnewitt
    Spectator

    If anyone is interested here is their web address
    http://www.fotona.si/lasers/dentin.html

    #7235 Reply

    Kenneth Luk
    Spectator

    Hi Eric,

    Great to read your speedy reply!

    Here’s the idea for using 7 mirror articulated arm:

    The articulated arm, is very user friendly. And it is the most efficient and long lasting delivery system available for mid infrared lasers. Unlike all of the other fibre and hollow-wave guide delivery systems it is permanent and the energy does not lose coherency and power travelling from the laser to the handpiece. It also allows the very short pulse of 80us to be delivered, which is difficult with a fibre. The fact that it is counter balanced and weightless means it does not need to be put in a holder every time you stop to see how the procedure is going, it just hangs in mid air ready to be used. Once dentists are shown how to use it properly they report that it is easy to use and in fact an advantage.

    To me ,it looked very bulky and not user friendly. Probably need to go and try it out for myself.

    What do you think?

    Hi Paul,

    So, you’ve already checked out on Fotona! 😉

    Are you changing your mind on Delight now ?

    I like the DElight very much! May be they can in-corporate VSP on their next generation of DElight ?

    I really missed not going to the Con-bio course. Not sure if Fotona runs comprehensive course for their customers.

    regards,

    Ken

    #7200 Reply

    ericbornstein
    Spectator

    Ken:

    I e-mailed Fotona to see where they are in the FDA process for my own info. I have yet to hear from them.

    When they get back to me I will let you know.

    Eric Bornstein DMD

    #7210 Reply

    Albodmd
    Spectator

    From the website I see they are based in Slovenia, are they working with a US distributor or doing it themselves. Interesting that they don’t list the Er:YAG as being used for desensitization of sensitive teeth, apthae and herpes, and Root canal sterilization, but do list those uses under the ND:YAG.
    Here’s a little info I found on the company:
    Fotona to Become Leading European Supplier of Medical Lasers

    Signing a contract on supply of fresh capital through government bonds in the value of SIT 700 million (almost 3.5 million euros), the Ljubljana-based laser producer Fotona is successfully concluding the second stage of its restructuring programme. The company left a zero net in the 1999 business year. It plans a 10-percent rise in sales in 2000. The sale of the military programme last year presented only nine percent of the overall sales.

    Fotona director Matjaz Lukac explained that the board would try to make Fotona the leading supplier of medical lasers in Europe. Fotona is an example of a company that has been successfully restructured with the help of the state, Lukac pointed out.

    Fotona was included in the government restructuring programme in 1998. After the latest supply of fresh capital the state will own 48.5 percent of the company, while the share of the Slovenian Development Corporation will total 16 percent.

    The introduction of new technology started the second stage of Fotona’s restructuring with the emphasis on the expansion of the civil programme. The concluding stage of the restructuring foresees disinvestment of excess capacity. The company intends to preserve the military programme, but will focus in future on the production of new medical and industrial lasers, which accounted for 55 percent of last year’s sales, followed by a 33-percent sale of optical communications.

    Fotona sold most of its products from the military programme to Sweden and Norway, while most of the optical communications were sold on the Russian and Ukrainian markets.

    #7189 Reply

    drnewitt
    Spectator

    Hi Ken

    No, I am pretty much sold on the DeLight at this point. Mainly based on the research I have done on here and talking to some laser owners. I have the DeLight rep coming by for lunch in about 15 minutes and the Biolase rep coming by tomorrow for lunch. No one from Font has offerd to buy me lunch yet smile.gif

    Don’t know enough about the history of Fotona with North American customers, support, training etc so it is not high on my list at this point.

    Looks like leasing is ready to go so its just a matter of signing on the dotted line. Nervous but excited.

    #7228 Reply

    Kenneth Luk
    Spectator

    Hi Al,
    Fontona also has a market share in Australia.

    Paul,
    In Hong Kong , I can get pretty good support with Conbio.
    Next to zero with Fotona except from Australia; pretty far away!
    DElight is still my choice; just need to go and rob a bank!

    Ken