Forums › Erbium Lasers › General Erbium Discussion › The Safety & Effectiveness of Dental Er:YAG Lasers – review

  • This topic is empty.
Viewing 6 posts - 1 through 6 (of 6 total)
  • Author
    Posts
  • #6924 Reply

    Kenneth Luk
    Spectator

    Hi Bob,
    Any chance of printing the article on this thread?
    I can’t get hold of one in HK (Hong Kong).
    Ken

    #6921 Reply

    drnewitt
    Spectator

    Our latest issue in Canada is August!

    #6922 Reply

    ericbornstein
    Spectator

    Kenneth:

    Here is a copy of the article that I had on my computer.

    I believe it is the same version that made it into the journal. Sorry for the length, but you asked for it.

    Safety and Effectiveness of Dental Er:YAG Lasers: A Literature Review With Specific Reference to Bone

    By Eric S. Bornstein, DMD and Mitchell A. Lomke, DDS

    A variety of laser wavelengths have been investigated for use in medicine and dentistry as realistic alternatives to conventional approaches to the cutting or removal of tissue during surgery. This literature review is intended to assess the potential of the dental infrared Er:YAG laser for safe and reliable thermal mechanical ablation of osseous tissues in the oral cavity, and to discuss the use of the Er:YAG laser for procedures involving the oral soft tissues and tooth enamel.

    Background

    In 1917 Albert Einstein published his theory of stimulated emissions of photons, and hence provided future scientists with the basis for creating a functional laser. (1) In 1960, Maiman created the first usable laser, which led to a search for possible medical applications for these devices.(2) Physicians and dentists have looked to lasers as an alternative to conventional mechanical instruments as a means of performing less invasive surgical procedures. In dental surgery, cutting of alveolar bone has been a major focus of research. To effectively cut alveolar bone without the thermal necrosis caused by other laser systems, a laser that selectively targets the chromophore of water and not the extracellular bony matrix is required.

    The electromagnetic energy emitted from a laser has a few unique characteristics that are fundamentally different from light energy evolving from a non-laser source. All of the photons emitted from a laser are coherent (moving in the same direction and phase), and all of the same wavelength (color). This allows a practitioner to exert maximum power per unit area irradiated, while simultaneously targeting a specific chromophore (tissue element) as an absorptive media for the unique wavelength of laser energy being used.
    The Er:YAG laser at a wavelength of 2.94µm has the highest specificity for water absorption of all the mid-infrared lasers.(3) In the past, this wavelength has been specifically used for very accurate ablation and cutting in different disciplines including ophthalmology , laryngeal surgery, and dermatology. Detailed histologic analyses of ablated tissue after irradiation with an Er:YAG laser have described well defined cuts with very small zones of thermal necrosis not greater than 40 µm (40/1000ths of a mm). In comparison, the thermal necrosis zones caused by Er:YSGG and CO2 laser systems were approximately twice as large, and histological samples observed with argon and Ho:YAG lasers produced thermal tissue necrosis zones of 200-300µm.(4) Therefore, among the major clinical advantages of using the Er:YAG laser is it’s ability to ablate both hard and soft tissues with minimal thermal damage.(5) Because of their unique absorption peak in water, Er:YAG generated photons warrant strong consideration as a useful tool to cut mineralized biologic tissues such as natural tooth structure and bone.

    Two important considerations for lasers that are highly specific for water absorption are thermal expansion and heat transfer through conduction. If these ideas are clearly understood, a clinician can maximize the desired effects while minimizing collateral tissue damage during laser irradiation.

    Er:YAG Interaction with Soft Tissue

    When using an Er:YAG laser in a pulsed mode with 200-400ms (micro-second) pulse widths, 5-40 microns of thermal mechanical tissue ablation (tissue removal), with as little as 5 microns of residual thermal damage, is generally observed for soft tissues. These characteristics result in a very narrow zone of thermal mechanical tissue ablation with minimal collateral thermal damage.(6)

    Venugopalan et al(7), described erbium tissue ablation in 1995. They surmised that to effectively cut or ablate human mucosa, the Er:YAG laser targets the chromophore of water selectively instead of the extracellular matrix of collagen. This produces an instantaneous vaporization of the water to a depth of about 4 µm per pulse. They observed that the mechanical integrity of the extra-cellular matrix (collagen) is not directly targeted when water is the dominant chromophore. Hence, to achieve tissue removal, the heated water (intracellular steam) expands, straining and later fracturing the matrix components (primarily collagen) in the extra-cellular environment.

    With the Er:YAG laser optical energy, the infrared beam is converted to local thermal energy in a confined tissue volume, and this occurs in the target chromophore of water. The result of this absorption in a thin layer of tissue is instantaneous tissue degradation or ablation. Accompanying this is an explosive ejection of the degraded cellular components and heated vaporous material.(8,9) Because Er:YAG energy has such a high absorption peak in water, the thermal damage to the tissue is kept to a minimum as the thermal mechanical ablation takes place (i.e. there is no charring). Improved healing of the irradiated tissues is the result. Neev et al.(10), in a thermo-optical skin conditioning study, observed that less thermal damage means less collagen remodeling is necessary. With less collagen damage and remodeling, more rapid healing with minimal scarring of soft tissue is seen.

    With the largest absorption coefficient in water of all the mid-infrared lasers, the Er:YAG has a very small optical penetration depth in water (less than one micron).(11) As a result, the tissue around the irradiated zone experiences only minimal damage, which is confined to 10 microns beyond the border of the tissue-beam interface. If the wavelength of the Er:YAG (2.94µm) is compared to conventional dental diode soft tissue lasers (810nm to 980nm wavelength), the depth of penetration per pulse with diodes is estimated to be greater by a factor of 104 (10,000x) or 4cm.(11) A diode laser is a solid-state laser with completely different chromophore targeting and absorption characteristics than the Er;YAG . Diode lasers have shorter wavelengths (800+ nm vs 2.94µm ), and thus have high absorption peaks in melanin and hemoglobin. This will cause the laser energy to pass through the water and produce thermal haemostatic and necrotic effects much deeper in the tissues as the photons are absorbed by the tissue pigments. Hence, the Er:YAG laser is safer and more controlled for purposes of cutting, but does not have good haemostatic properties. (11) Procedures such as a gingivectomy, gingivoplasty, frenectomy, incision and drainage, and removal of a fibroma are easy to accomplish using the Er:YAG laser, but the
    haemostatic properties are only marginal. (12)

    Erbium laser effects on teeth

    In 1997, the US Food and Drug Administration approved the Er:YAG laser for caries removal, cavity preparation, and laser etching of enamel.(13) This is the most effective laser in the mid-infrared spectrum for the ablation of dental enamel, and the photobiology of the Er:YAG-enamel interaction has been defined.(9,10,14) Water molecules within the prismatic enamel layer of a tooth make up only 4% of its chemical composition, however, represent 11% of enamel’ s total volume. As the laser energy interacts with the enamel matrix, water absorbs the laser energy and rapid vaporization of the water occurs. As this photothermal reaction takes place, the steam generated within the enamel is associated with a volumetric expansion within the enamel matrix. This in turn produces microevaporative explosions that result in a thermally driven, mechanical ablation of the tooth structure.(14)

    The effect of water spray and thermal tissue effects with the Er:YAG laser on both hard tissues and dental materials was investigated and described by Wigdor, Visuri and Walsh.(15,16) They found that as long as enamel is continually hydrated with a water spray during the laser exposure (serving as a heat sink) , there will be only a minimal increase in temperature of the enamel layers and deeper tissues. These data are clinically important. Friedberg et al(17) reported on another FDA approved laser for cutting of teeth, the Er:YSGG. They found that additional energy is required from the unit to initiate enamel ablation when the laser is propagated through an aerated mixture of water droplets. This method was found to attenuate the beam, as opposed to the water being sprayed directly on the irradiated area of enamel and acting as a heat sink.(17)

    Bactericidal Effects of the Er:YAG Laser

    A major advantage of the Er:YAG laser as compared to mechanical therapy for procedures on biologic tissues is the bactericidal effect of the laser. During the Er:YAG tissue interaction, the bacteria in the path of the beam are destroyed as the water within the bacterial cells undergoes the same instantaneous phase change (liquid to steam) as the water in the tissue matrix being ablated.(18,19) The destruction of bacteria during treatment may be an important ancillary effect of the use of the laser. A study by Ando et al.,(18) demonstrated the pronounced potential of the Er:YAG laser on periodontal bacteria even at low energy levels.

    Current Medical Applications of Er:YAG Lasers for Bone

    Er:YAG bony ablation has been successfully accomplished in a variety of clinical situations. In a review of the use of this laser for middle ear ablation procedures, Frenz et al (20) stated that otological surgeons have identified the advantages of this laser for osseous surgery. These are the water absorption characteristics of the wavelength, and the precise bone ablation with minimal residual thermal energy. The histological data for Er:YAG lasers, when used by otological surgeons to perforate stapes footplate (osseous ear tissue) in surgical procedures, demonstrates that the lateral thermal damage associated with the bony cut was restricted to only 5-10µm beyond the cuts.

    When attempting to cut nasal bone with a laser, Truong et al (21) described two specific criteria that should be met before a laser wavelength is chosen. These are rapid tissue ablation and an absence of visible char in the ablation site. They concluded that this laser produced an excellent result, and further stated that the addition of a water spray on the area being irradiated during ablation resulted in the lowest level of char and presumably the lowest thermal damage to the tissue. It was also noted that as long as the surface of the ablated tissue remained moist with a water spray throughout the ablation process, and the laser was used at appropriate settings, char was not seen regardless of the size/volume of the treated tissue.(21) These results agree with an earlier study by Romano et al(22) using Er:YAG lasers to cut cortical bone. These investigators observed a linear increase of ablation depth with the number of pulses. A greater number of pulses, translates into more ablative energy in the area being irradiated per unit time. It was found that as long as adequate water spray was provided to the area, more bone was ablated with additional pulses (per unit time) without the char formation that would normally be an expected consequence of extra heat from the added pulses. It was concluded that repetition rates up to tens of Hz did not significantly increase the peripheral thermal damage to the bone. (Hertz [Hz] is the SI unit of frequency for electromagnetic radiation per second [10 Hz equals 10 laser pulses per second]). These studies are germane since all commercially available FDA approved Er:YAG dental lasers function from 7 to 30Hz with a continuous water spray option.
    In an attempt to quantify and illuminate the important thermal interactions of erbium lasers with calcified biologic tissues, it is important to review the inherent quantum mechanics and spectroscopy characteristics of the erbium laser/water interaction. In a study conducted by Shori et al,(23) describing the quantum mechanics of the water molecule under erbium irradiation, it was demonstrated that as water absorbs more of the incident energy of an Er:YAG laser, and the temperature of the water increases, the length and strength of the oxygen-hydrogen bonds in the water molecule changes because of the large increase in kinetic energy . In addition, as this lengthening and weakening of the bonds occurs , the absorption peak for the water molecule shifts to wavelengths that are significantly shorter than 2.94um. T his negative shift in the absorption peak for water greatly diminishes the effectiveness of the beam to perform controlled thermal mechanical ablation on tissues.

    The findings of Shori et al(23) have important implications for how erbium lasers interact clinically with calcified biologic tissues. If more heat than is necessary to induce ablation is delivered to tissue , the negative absorption coefficient shift in water will be larger and more profound. This quantum, spectroscopic, and thermal reality associated with erbium lasers needs to be clearly understood by the clinician contemplating osseous surgery with an erbium laser. To prevent excess thermal deposition, and a less effective thermal mechanical ablation, the clinician should keep a continuous water spray on the irradiated area to act as a heat sink, and use the least amount of energy necessary to avoid the occurrence of the negative absorption shift.

    Recently, Sasaki et al(24) analyzed the effects of bone removal by the Er:YAG laser under scanning electron microscopy and infrared spectroscopy. They concluded that the Er:YAG laser ablation of bone produced a groove with similar dimensions to that of a drilling bur. In addition, the chemical composition of the bone surface after Er:YAG ablation was much the same as that following bur drilling. Furthermore, the production of toxic substances was not observed following Er:YAG laser irradiation or conventional bur drilling of bone. Their conclusion was that the use of Er:YAG laser ablation may represent an alternative approach for oral and periodontal osseous surgery.

    Conclusion

    The literature pertaining to the potential use of the dental Er:YAG laser for osseous surgery, as well as am overview of current clinical use for soft tissue and tooth enamel applications, has been reviewed.

    Presently in the United States, there is only one laser wavelength that is FDA approved for intraoral osseous surgery , the Er:YSGG laser (which produces photons at 2.79µm). (25) Eversole and Rizoiu, describing the cutting of cortical rabbit tibia with this hydrokinetic Er,Cr:YSGG laser system, showed that when this laser system was used at 140 microseconds and 20Hz, 24 hrs post exposure photomicrographs of the wound cavities presented histology characteristic of a thermal coagulative effect,with a zone measuring 40-60 microns, similar to that of a dental bur. (26)

    In this review of the literature, the data examining the photobiology of the Er:YAG laser supports its use in intraoral osseous surgery. As discussed in this review of the literature, if the criteria of rapid and char-free tissue ablation (21) and excellent water absorption characteristics with precise bone cutting (20) are considered, the Er:YAG laser should be considered safe for intraoral osseous surgery.

    Dr. Bornstein graduated from Tufts University School of Dental Medicine in 1992, and the Maimonides Medical Center General Practice Residency program in Brooklyn, NY in 1993. He practices general, implant and laser dentistry in Natick, Massachusetts, and operates the Metrowest Maxillofacial Imaging Center at the same location. Disclosure: Dr. Bornstein is a consultant for OpusDent USA concerning matters of Photobiology and Laser T issue Thermodynamics. He owns two combination Er:YAG/CO2 dual wavelength lasers and two 830nm diode lasers all purchased from Opusdent USA.

    Dr. Lomke graduated from the University of Maryland, Baltimore College of Dental Surgery in 1979, and a US Public Health Service General Practice Residency in 1980. He practices general, laser, and aesthetic dentistry in Silver Spring, Maryland. Dr. Lomke has achieved the level of Educator Status with the Academy of Laser Dentistry, with Advanced Proficiency in Erbium and Diode wavelengths. Disclosure: Dr. Lomke is a clinical instructor for OpusDent USA. He owns one dual wavelength Er:YAG/CO2 laser and one 830nm diode laser both purchased from OpusDent USA.

    References

    1.Einstein A (1917) Zur Quantentherorie der Strahlung. Physiol Z 18:121-128.

    2.Maiman T (1960) Stimulated optical radiation in ruby. Nature 6:493-494

    3.Hale, G.M., Querry, M.R. (1973): Optical constants of water in the 200-nm to 200-um wavelength region. Appl. Opt. 12, 555-563

    4.Gerber, B: Lasers in the Musculoskeletal System. Springer, Berlin pp 11-17

    5.Aoki, A. et al, Er:YAG Clinical Experience in Japan: Review of Scientific Investigations. SPIE Vol. 3248, 1998

    6.Lanigan, S: Lasers in Dermatology, London, Springer, pp 57-79, 2002

    7. Venugopalan V, et al: Pulsed Laser Ablation of Tissue: Surface Vaporization or Thermal Explosion?, Wellman Laboratories of Photomedicine, Harvard Medical School, SPIE Vol. 2391, 1995.

    8. Walsh J, et al: Er:YAG Laser Ablation of Tissue: Effect of Pulse Duration and Tissue Type on Thermal Damage, Lasers in Surgery and Medicine, 9:314-326,1989

    9 . Walsh J and Deutsch F, Er:YAG Ablation of Tissue: Measurement of Ablation Rates, Lasers in Surgery and Medicine, 9:327-337,1989.

    10. Neev J, et al: Thermo-optical Skin Conditioning: A new method for thermally modifying skin conditions. Lasers in Surgery: Advanced Characterization, Therapeutics, and Systems XII, SPIE Vol 4609, 2002

    11 Niemz, M: Laser-Tissue Interactions. Fundamentals and Applications, Berlin, Springer, pp45-80, 2002

    12 Kesler, G. et al: “Periodontal plastic surgery” – thermal effect analysis using Erbium:YAG Kesler’s handpiece. Lasers in Dentistry VI, SPIE Vol 3910, 2000

    13. Lasers in Periodontics, Academy report. J. Periodontol 2002; 73: 1231-1239.

    14 Moshonov J, Stabholz A, Leopold Y, et al: Lasers in dentistry. Part B— Interaction with biological tissues and the effect on the soft tissues of the oral cavity, the hard tissues of the tooth and the dental pulp. Refuat Hapeh Vehashinayim 18(3-4):21-28, 107-8, 2001.

    15. Wigdor, HA, Visuri, SR, Walsh, JT. Effect of water on dental material ablation of the Er;YAG laser. SPIE Proc 1994

    16. Visuri, SR, Walsh JT, Wigdor, HA. Erbium laser ablation of hard tissue: Modeling and Control of the thermal load. SPIE Proc 1994;2128

    17 Freidberg et al., Pulsed erbium laser ablation of hard dental tissue: The effects of atomized water spray vs water surface film,Lasers in Dentistry VIII, SPIE Vol.
    4610 (2002)

    18 . Ando Y, et al: Bactericidal effect of Er:YAG laser on periodontopathic bacteria. Lasers Surg Med 19:190-200, 1996.

    19. Clayman L and Kuo P, Lasers in Maxillofacial Surgery and Dentistry, New York, Thieme, pp. 19-28.

    20 Frenz, M. et al, Laser Applications in Middle Ear Surgery: Advantages and Possible Side-Effects. Lasers in Surgery: Advanced Characterization,Therapeutics, and Systems XII, SPIE Vol. 4609, 2002

    21. Truong, M. et al, Erbium:YAG Laser Contouring of the Nasal Dorsum: A Preliminary Investigation. Lasers in Surgery: Advanced Characterization. Therapeutics, and Systems XI, SPIE Vol. 4244, 2001

    22 Romano, V. et al, Bone Microsurgery with IR-Lasers: a comparative study of the thermal action at different wavelengths. SPIE Vol. 2077, 1984

    23 Shori et al, Quantification and Modeling of the Dynamic Changes in the Absorption Coefficient of Water at 2.94um. Journal on Selected Topics in Quantum Electronics, Vol 7, No 6, Nov/Dec 2001 pp959-970.

    24 Sasaki KM. et al, Scanning Electron Microscopy and Fourier Transformed Infrared Spectroscopy Analysis of Bone Removal using Er:YAG and CO2 Lasers. J Periodontol, 2002 Jun;73(6):643-52

    25 Food and Drug Administration. 510k Summary of Safety and Effectiveness Information: Waterlase Millenium, hydrokinetic tissue cutting system. K013908

    26 Eversole L.R., Rizoiu I, Kimmel A., Osseous repair subsequent to surgery with an erbium hydrokinetic laser system, International Laser Congress, International Proceedings Division, Sept 25-28, Athens Greece 1996, pp 213-221

    #6926 Reply

    Swpmn
    Spectator

    Eric:

    Great summary with excellent literature review!!! The points made regarding use of copious water irrigation to act as a “heat sink” for hard tissue ablation safety AND marginal soft tissue hemostasis with erbiums correlate strongly with what I observe clinically. Looking forward to receipt of the Dentistry Today article.

    Al

    #6925 Reply

    Kenneth Luk
    Spectator

    Eric,

    Thanks for posting the article! Great review!

    BTW did Fotona gave you anymore info about their Erbium?

    Ken

    #6923 Reply

    ericbornstein
    Spectator

    Al and Ken:

    Thank you for the kind words about the article. Mitch and I put many hours and(Columbia Dental School peer-review mandated) re-writes into it.

    I hope it gives newer laser practitioners a clearer perspective on the energies, and what they can and cannot accomplish.

    Ken, I did e-mail Fotona, and they have yet to get back to me.

    Eric Bornstein DMD

Viewing 6 posts - 1 through 6 (of 6 total)
Reply To: Reply #6924 in The Safety & Effectiveness of Dental Er:YAG Lasers – review
Your information: