• This topic is empty.
Viewing 11 posts - 1 through 11 (of 11 total)
  • Author
    Posts
  • #11452 Reply

    Lee Allen
    Spectator

    Bravo Graeme!

    The photos are great, too.

    I have a patient that wants the laser used always and needs a crown and veneer. What settings would you suggest?

    And with the “spray” interproximal affecting the soft tissue, would you think it advisable to sheild the papilla with a bladed instrument?

    My guess at settings is 5.5 Watts for the G-6 bulk reduction enamel and dentin, and 2.75 W for the marginal refinement with the C-6.

    I get some dental charring over 3.5W which must be the high vacuum placed too close even with 30% Water?

    Nice results.

    #11449 Reply

    Graeme Milicich
    Spectator

    Lee
    You are right on with your settings.
    You can use some sort of isolation if you so desire. I still tend to place cord if I have any issues. It only takes a few seconds.

    Re the charring of dentin. The issue here is one of energy density (which is altered by simply defocusing) The problem with dentin is it is made up of two components, hydroxyapatite and collagen. Each react differently to the same level of laser energy, and there si nothing we can do about the physics of this. I just ensure I finish off in a defocused mode.

    Out of interest, the photos are taken with the ultracompact pocket Canon Ixus 500 (about as small as they get. I can operate it using one hand) Because it is so small, the flash is right beside the lens so you don’t get any shadows. I shoot in macro mode while sitting behind the patient. No fancy setup. Just point and squirt. The 5million pixel image then lets you crop it down to what you want, and still maintain good quality images
    Cheers

    #11443 Reply

    2thlaser
    Spectator

    Hi Lee,

    I am still in the Czech Republic, leaving to come home tomorrow. I would advise that once you get your margins to where you kind of want them, lower your power settings to 1.5W or less. Final marginal refinement should be done at .5-.75W, especially if you are using magnification, you will see that the C-6 tip will cut VERY VERY effectively at .5W. Sometimes I might even place my margins at that setting, but reduce the water to 15% or so, and air to 20% or so, depending on your visualization of your surgical area. I get home tomorrow night, and will be at work on Wednesday and Thursday before heading to Las Vegas for a Barr/Colonna training course, please feel free to call!

    All the best,
    Mark

    (Edited by 2thlaser at 7:05 pm on Nov. 1, 2004)

    #11446 Reply

    Glenn van As
    Spectator

    Neat case Graeme and yes the internet has its hosts of people intent on stealing ideas and photos as their own.

    Graeme neat case and nice photos. I have found that all my laser preps have a V type notch. Is this not so for the Waterlase. I can show you cuts on root with a 400 micron tip which is easy to demonstrate. Have you noticed different shaped cuts with various tips.

    Do smaller tips behave differently?

    THanks for a neat case……

    Glenn

    #11447 Reply

    Graeme Milicich
    Spectator

    Glen
    The Waterlase has got a new tip that has quite a different energy footprint that allows you to deal with certain phases of soft tissue and cavity preparation in a different way.
    Some of the effects you can get with the C tip are quite cool
    Cheers

    #11450 Reply

    Graeme Milicich
    Spectator
    QUOTE
    Quote: from Glenn van As on 4:55 am on Nov. 2, 2004
    Neat case Graeme and yes the internet has its hosts of people intent on stealing ideas and photos as their own.

    Graeme neat case and nice photos.  I have found that all my laser preps have a V type notch.  Is this not so for the Waterlase.  I can show you cuts on root with a 400 micron tip which is easy to demonstrate.  Have you noticed different shaped cuts with various tips.  

    Do smaller tips behave differently?

    THanks for a neat case……

    Glenn

    Gleen
    I posted another case that showed the adaptability of the C6 tip
    http://www.rwebstudio.com/cgi-bin….start=0
    You can tune how it cuts by focusing and defocusing, which changes the energy distribution pattern over the footprint of the beam as you move the tip in and out of focus
    Cheers

    #11445 Reply

    Hi Graeme ……..I saw this case and thought it was neatly handled.

    Is there a picture on the Biolase web site of the C6 tip….I am off to look at it.

    I wonder if having 30 Hz makes a difference compared to 20 Hz for smoothness of cut (it does on my laser on both hard and soft tissue).

    WHich begs another question is with the new MD by waterlase. What is the maximum Hz rate for hard tissue. IS it 50 Hz and then of course 160 mj for 8 watts or does the maximum Hz setting of 50 hz only apply for soft tissue where it will make a big difference in cutting smoothness ( I see this in varying between 10 Hz and 30 hz on my machine).

    One issue with 50 hz for hard tissue is if the relaxation time between pulses is enough to continue doing the same amount of restorative work without anestetic.

    I know that the pulse duration is short (good stuff) but I just wonder about that.

    Graeme, I have worked with the 400 micron , 600 micron and chisel tips and even a soft tissue tip of 200 microns with flooding from a high speed syringe to cut preps. I have seen all of them cut a V notch especially in dentin.

    I might try that and take a picture for everyone comparing the cuts at 5 – 10 seconds at high mag on cementum once I get back in the office. I can tell you the smaller tips cut far far faster but the shape of the cut was the same.

    Off to see what the C 6 tip looks like.

    Cya

    Good stuff

    Glenn

    #11442 Reply

    Graeme: great stuff! And to think, 5 years ago we didn’t think crowns could be done with lasers! Nice job raising the bar.

    Kelly

    #11451 Reply

    Graeme Milicich
    Spectator
    QUOTE
    Quote: from kellyjblodgettdmd on 8:18 am on Nov. 4, 2004
    Graeme:  great stuff!  And to think, 5 years ago we didn’t think crowns could be done with lasers!  Nice job raising the bar.

    Kelly

    Mark raised the bar, we all followed in his footsteps.

    Once you understand how to create smooth cuts with the laser, anything is possible with a bit of practice.

    Cheers

    #11448 Reply

    Graeme Milicich
    Spectator
    QUOTE
    Quote: from Glenn on the road on 7:09 am on Nov. 3, 2004

    One issue with 50 hz for hard tissue is if the relaxation time between pulses is enough to continue doing the same amount of restorative work without anestetic.  

    Cya

    Good stuff

    Glenn

    Interesting thought.
    Apparently (according to research conducted at Queensland University by Prof Lawrie Walsh) the A and C fibres have a natural frequency of 15-20Hz which is why a pulsed laser running at that frequency has the abliity to create analgesia (not anaesthesia).So in theory, a multiple of these frequencies might have a similar, greater or lesser effect.

    Not having a variable pulse laser, I can’t play with this to find out.
    This will be up to all the lucky ones with the new MD laser to sort out.

    Cheers

    #11444 Reply

    2thlaser
    Spectator

    I think we forget that as we increase the pulse rate, the depth of penetration decreases drastically. From what I have seen, at 50hz, one doesn’t get anywhere near the depth of cut you get at 30hz or below. I think that because of this, it IS just as easy to get and maintain good laser anesthesia, even at 50hz. Again, I have had the luxury of variable hz on my Waterlase for over 2 years now, and Graeme, I have been doing more at 15hz, per our conversation, with GREAT results so far, using far less power obviously. Just some observations…..again, tip selection is another VERY important consideration in this equation due to the energy pattern one needs to use to do their work.

    Mark

Viewing 11 posts - 1 through 11 (of 11 total)
Reply To: Reply #11444 in First Crown Prep
Your information: