Forums › Diode Lasers › General Diode Forum › Laser Perio

  • This topic is empty.
Viewing 15 posts - 31 through 45 (of 64 total)
  • Author
    Posts
  • #8156 Reply

    whitertth
    Spectator

    Boy, I am I really loving I started this thread…….

    #8167 Reply

    Glenn van As
    Spectator

    Now Bob and all , this thread is important for understanding why certain wavelengths may not be the best for periodontal therapy as well.

    My suggestion has always been that a one laser for all mentality is not a fair assessment of the benefit of lasers for dentistry.

    I have learned alot on this thread as well……..it deserves a little gold star or something as a cant miss thread.

    Thanks Bob et al for the great read.

    Glenn

    #8165 Reply

    Dan Melker
    Spectator

    When analyzing bone regeneration one must consider certain facts-
    1. 5 year followup of furcation procedures for bone regeneration showed significant failure rate occurring. Vertical regeneration occurred but not regeneration to the furcation itself.
    2. I have had several radiographic successes only to open up the area and find failure.
    3. In 1980 I did a video with Dr. D. Walter Cohen and Dr. Burton Langer on bone regeneration using Hydroxyappetite. I showed extremely succesful radiographs only to do re-entry and find fibrous encapsulation of material-no regeneration though no probing or bleeding upon probing.
    4. Root irregularities are more important than we realize for success of regeneration procedures.
    5. Kelly asked why open up cases? Not laser technique. That will really be the only way to determine success. X-rays can always be misleading.
    6. I am so impressed with Bob’s material and am sure regeneration will be found in re-entry.
    7. Personally I think if Ron reentered his case he would not be so happy. As I said before, I think the angulation of the x-ray is very misleading.
    This is not a negative take on what I see-this is an objective statement, based on what we all know to be true about x-rays-they are misleading. Visual observation solves the x-ray problem-no one  can deny what they see.
    Root form is critical!
    Danny

    (Edited by Dan Melker at 8:52 pm on June 6, 2004)

    #8171 Reply

    Kenneth Luk
    Spectator

    Bob,

    How does one render granulation tissue aseptic? Pulsed 1064nm near-infrared that penetrates 2-4 mm into soft tissue and kills anaerobic bacteria (P Gingivalis, Intermedius, Forsythus) and other pathologic proteins (endotoxins, PGE2, MMPs, etc) without burning or vaporizing the granulation tissue, capillaries, or progenetor cells themselves as they would require much more energy to effect than the path proteins).

    Whta is the difference between 1064 and diodes ( 810,980) in tissue penetration / bacterial decontamination?
    Can diode render granulation tissue aseptic?

    Ken

    #8187 Reply

    Robert Gregg DDS
    Spectator

    Hi Ken,

    Thanks for the email!

    Here’s part of the answer to your question in a study done by Dr. David Harris:

    0855 Ablation of Porphyromonas gingivalis in vitro with Pulsed Dental Lasers
    D.M. HARRIS, University of California, San Francisco, USA

    Abstract  

    Both pulsed Nd:YAG (1064nm)and continuous diode (810nm) dental lasers are in current use for treatment of periodontitis. It has been shown that laser treatment kills pathogenic bacteria (laser antisepsis), but a quantitative method for determining clinical dosimetry does not exist.

    Objective: The purpose of this study was to develop a method to quantify the efficacy of ablation of Porphyromonas gingivalis (Pg) in vitro for two different lasers.

    Methods: The ablation thresholds for the two lasers were compared in the following manner. The energy density was measured as a function of distance from the output of the fiber-optic delivery system. Pg cultures were grown on blood agar plates under standard anaerobic conditions. Blood agar provides an approximation of gingival tissue for the wavelengths tested in having hemoglobin as a primary absorber. Single pulses of laser energy were delivered to Pg colonies and the energy density was increased until the appearance of a small plume was observed coincident with a laser pulse. The energy density at this point defines the ablation threshhold.

    Results: Ablation thresholds to a single pulse were determined for both Pg and for blood agar alone. The large difference in ablation thresholds between the pigmented pathogen and the host matrix for pulsed-Nd:YAG represented a significant therapeutic ratio and Pg was ablated without visible effect on the blood agar. At threshold the 810-nm diode laser destroyed both the pathogen and the gel.

    Conclusion: Clinically, the pulsed Nd:YAG may selectively destroyed pigmented pathogens leaving the surrounding tissue intact. The 810-nm diode laser may not demonstrate this selectivity due to its greater absorption by hemoglobin.

    <a href="http://iadr.confex.com/iadr/2003SanAnton/techprogram/abstract_27983.htm

    Millennium” target=”_blank”>http://iadr.confex.com/iadr&#8230;.lennium did not fund this study.

    The reason for the difference in tissue effects and selectivity has to do with peak powers not wavelength.

    The free-running nature of pulsed Nd:YAGs is what delivers near-infrared (N-IR) energy to the pigmented pathogens with enough intensity to penetrate the mass of granulation tissue and denature sensitive proteins while those not as highly absorbed by the N-IR wavelength (which is still a factor) are left viable.

    If one could pulse a N-IR diode in the millionths of a second, and allow for high peak powers and long thermal relaxation times, then it ought to work as well.

    Does that help?

    Bob

    #8168 Reply

    Kenneth Luk
    Spectator

    Hi Bob,
    Does this mean due to the lower haemoglobin absorption by 1064, a more regenerative clot can be produced even if a diode can achieve pulse width in micro seconds. I believe there is one 985nm by Vision(?) that can achieve 150 micro sec.
    At what depth is tissue penetration to with 810 and 980?
    Thanks,
    Ken

    #8179 Reply

    Robert Gregg DDS
    Spectator
    QUOTE
    Quote: from Kenneth Luk on 11:37 pm on June 29, 2004
    Hi Bob,
    Does this mean due to the lower haemoglobin absorption by 1064, a more regenerative clot can be produced even if a diode can achieve pulse width in micro seconds. I believe there is one 985nm by Vision(?) that can achieve 150 micro sec.
    At what depth is tissue penetration to with 810 and 980?
    Thanks,
    Ken

    Hi Ken,

    Since I have never heard of or seen a FR pulsed diode, I can’t answer as to whether there would be more or less absorbtion in hemoglobin as to make much difference.

    There is no tissue penetration into perio pocket tissues. That is the point of Harris’ study. CW diodes only have a surface effect as they are most often used in a “hot tip” or “activated” or “conditioned” fiber mode.

    FRP Nd:YAGs are never used in an activated or conditioned mode in the perio pocket. That allows for the beam to transmitt into the tissues while keeping intensity and selectivity at depth.

    I have a great demo/parlor trick for this. I can vaporize completely the black ink through an anesthetic carpule filled with water/anesthetic. When a diode is used, the plastic wrapper ignites on fire…..

    Bob

    #8159 Reply

    Dan Melker
    Spectator

    Bob,
    I am correct-is the Periolase the only laser that creates a fibrin clot which appears so critical for bone regeneration. This is not a loaded question. D. Kimmel and I have been having great discussions about different lasers.
    Thanks,
    Danny

    #8194 Reply

    Robert Gregg DDS
    Spectator
    QUOTE
    Quote: from Dan Melker on 9:27 am on June 30, 2004
    Bob,
    I am correct-is the Periolase the only laser that creates a fibrin clot which appears so critical for bone regeneration. This is not a loaded question. D. Kimmel and I have been having great discussions about different lasers.
    Thanks,
    Danny

    Morning Danny,

    No, there are other FRP Nd:YAGs out there that have “long” pulse durations and can create the desired hemostatic effect.

    Lares Research is presently the only other company in the USA that has a “short” PD of 150 usec and a single “long” pulse duration of 635 usec that Del and I disclosed to them when they made their laser for MDT to resell.  They use a different nomenclature now at our insistance of 110 and 280 or something so as not to confuse their parameters with ours.  (We think our confidentiality agreement they signed prevents their resale of a “dual pulse” to the dental market, but that’s a lawsuit for another time).

    The PerioLase MVP-7 has 100, 150, 250, 350, 450, 550, and 650 usec pulse durations.

    100 is good for cutting with minimal thermal injury (due to short “on” time).  Useless for hemostasis.  Highest peak powers = deepest penetration and bug kill rates.

    150 is good for cutting with a little hemostasis.  A nice “tweener” for most clinical situations.

    250 – 650 is all about using the appropriate amount of PD to control the hemostatic repsonse depending on the clinical situation, and that allows the clincian to select the rate of speed they want to move.  Nice to have a laser that is responsive to the operator and not the other way around.

    At 650 one needs to move pretty fast or have a massive bleeder.  At 350 one can move the fiber less rapidly w/o causing a problem.

    700 usec is the threshold for human tissue and thermal relaxation (aka thermal relaxation constant).

    Hence the reason we backed away from a laser with 700 or more.  We used a medical laser with 700 usec and up to 1000 usec and 700 and above is just to hard to use and avoid the damaging effects to collateral tissues.:o  Sure it controls bleeding!  But the thermal injury potential is just to great and happens quickly for even a skilled FRP Nd:YAG laser operator.

    So unless the thermal properties of human tissues change anytime soon, the Periolase MVP-7 operating parameters will not be changing!  As a result there are no plans to “improve” on the PerioLase.  Maybe a color screen someday???

    There are other FP Nd:YAGs that have “long” pulse durations, but they are not available in the USA–like the Fotona Fidelis.  I think they have 150, 350 and 1000 usec.  Interesting selection–no doubt made by an engineer–but they don’t make a lot of clinical sense to me.  If I had only 3 PDs to choose from, those would not be the ones.

    The question becomes which PD to use under what conditions, how, and at what power and repetition rates–hence the emphasis with MDT on clinical training (not a CD-ROM with “borrowed” and incomplete info).  There are few clincians with experience with various long PDs in tissue with Nd:YAG–and most (in fact all in the USA) are using a laser parameters of our design and configuration.

    I spoke with C.C. Rice from Texas yesterday.  He’s a long time single pulse (100usec) Nd:YAG user.  Been using the PerioLase MVP-7 for about 3 months now.  He just laughed at the ease of hemorrhage control compared to his single pulse device.  He was calling the Periolase a “turbo” “four-wheel drive” laser for it ability to adjust to the different clinical conditions.  I thought that was a pretty good description.

    Thanks for the great question!

    Bob

    #8203 Reply

    Swpmn
    Spectator
    QUOTE
    There is no tissue penetration into perio pocket tissues.(WITH DIODES, RIGHT BOB?).  That is the point of Harris’ study.  CW diodes only have a surface effect as they are most often used in a “hot tip” or “activated” or “conditioned” fiber mode.

    FRP Nd:YAGs are never used in an activated or conditioned mode in the perio pocket.  That allows for the beam to transmitt into the tissues while keeping intensity and selectivity at depth.

    Bob

    So this is a major difference in how diodes and Nd:YAGs are used in periodontal pockets. How come a diode couldn’t be used un-activated in a perio pocket? Is it because we can’t pulse the diode and so it has to be kept power-weak to avoid thermal damage? The mechanism of action with the diode in a perio pocket is to apply heat with an activated fiber so that perio pathogens are killed?

    But as the Nd:YAG is used in the pocket, won’t a buildup of blood, or dead pigmented pathogens or pigmented tissues eventually “initiate” the fiber tip so that it does begin to lean toward the “hot glass effect”?

    With LPT, do you wipe the tip of the fiber frequently or keep re-cleaving the fiber tip to reduce a shift towards an activated fiber tip?

    Al

    #8201 Reply

    Swpmn
    Spectator
    QUOTE
    At 650 one needs to move pretty fast or have a massive bleeder.

    Bob

    I’m confused by this statement and need clarification.

    Do you mean?:

    1) In most instances, if the practitioner is using the Periolase at 650 usec PD he/she needs to move the fiber tip rapidly to avoid tissue damage(because the laser is firing for a long time at 650 usec)

    2) However, the 650 usec setting can also be used in a focused state when the practitioner has a large blood vessel with difficult to control hemorrhage

    Thanks,

    Al

    #8202 Reply

    Swpmn
    Spectator

    One more question for anyone that has used lasers for treatment of periodontitis. Gregg, Kaminer, Kimmel, Schalter please chime in here:

    I don’t know much about either diode or Nd:YAG periodontitis treatment protocols but it seems I’ve been told the diode fiber should be placed 1mm shorter than the perio pocket. On the other hand, I think Rusty Schaeffer told me with the LPT protocol the fiber should be placed to the full depth of the periodontal pocket???

    What is the logic behind these differences? Does it relate to activated vs. inactivated tips or pulsed, high-peak powers attainable with the Nd:YAG?

    If the doctor has to re-measure each perio pocket and then re-set his fiber length with a diode, isn’t that a real PIA(Pain in the A&#36&#36)? Who actually does this and how practical is it considering diode laser handpiece design? We should have fibers that are marked in 1mm increments!!! But the laser would probably focus on any pigmented markings.

    So on number four I have a mesio-buccal 4mm and on the disto-buccal I have a 6mm. I’m actually going to stop and extend my diode fiber from 3 to 5mm for the DB pocket?rock.gif? This seems silly and not clinically reasonable.

    Al

    #8141 Reply

    Anonymous
    Guest

    I’ve been told the diode fiber should be placed 1mm shorter than the perio pocket.

    True

    On the other hand, I think Rusty Schaeffer told me with the LPT protocol the fiber should be placed to the full depth of the periodontal pocket???

    True

    Does it relate to activated vs. inactivated tips or pulsed, high-peak powers attainable with the Nd:YAG

    I think different objectives, part of the FRVPnd:YAGg objective is to gain access, remove debris,then stimulate osteoblasts and create a clot allowing the body to heal. The diode objective is to remove inflamed tissue and prevent epithelial downgrowth.

    If the doctor has to re-measure each perio pocket and then re-set his fiber length with a diode, isn’t that a real PIA(Pain in the A&#36&#36)?

    IMO Yes

    Who actually does this and how practical is it considering diode laser handpiece design?  We should have fibers that are marked in 1mm increments!!!  But the laser would probably focus on any pigmented markings.

    Unless Tx protocols have changed, don’t forget each subsequent Tx is 1 mm shorter so your 6mm pocket(below) will require 3 visits sad.gif

    So on number four I have a mesio-buccal 4mm and on the disto-buccal I have a 6mm.  I’m actually going to stop and extend my diode fiber from 3 to 5mm for the DB pocket?rock.gif?  This seems silly and not clinically reasonable.

    Agreed

    #8180 Reply

    Robert Gregg DDS
    Spectator
    QUOTE
    Quote: from Swpmn on 6:57 pm on June 30, 2004

    Quote:
    There is no tissue penetration into perio pocket tissues.(WITH DIODES, RIGHT BOB?).

    Right

    Right you are above Al. And………

    I’ll answer your questions in a Q&A format below:

    So this is a major difference in how diodes and Nd:YAGs are used in periodontal pockets.

    Yes

    How come a diode couldn’t be used un-activated in a perio pocket? Is it because we can’t pulse the diode and so it has to be kept power-weak to avoid thermal damage?

    Exactly And with the power down, there isn’t sufficient beam intensity for penetration for bug kill. With the power up, there is too much collateral damage.

    The mechanism of action with the diode in a perio pocket is to apply heat with an activated fiber so that perio pathogens are killed?

    Correct

    But as the Nd:YAG is used in the pocket, won’t a buildup of blood, or dead pigmented pathogens or pigmented tissues eventually “initiate” the fiber tip so that it does begin to lean toward the “hot glass effect”?

    Brilliant deduction! That makes perfect logic and sense. And to a degree it does occur. But not to the degree it over powers the physics of high peak powers.

    But this brings up a good point and a common misconception for those who have been using FRP Nd:YAGs over the years and/or diodes.

    There is a difference between the hot glass effect where the fiber is intentionally “initiated” in order to block the exit of photons in the tip and cause incandescence since the tip is around 800 degrees Celcius; and the hot tip effect where proteins build up on the edge of the fiber, but the forward beam penetration is still very intense. Furthermore, the protein hot tip effect is somewhere around 100 degrees Celcius versus 800 degrees.

    With LPT, do you wipe the tip of the fiber frequently or keep re-cleaving the fiber tip to reduce a shift towards an activated fiber tip?

    I wipe the fiber off because the accumulated debris blocks my entry into to the pockets. I actually like a little of the “side-heating” effect from the hot protein tip effect since it “pulls” the sulcular debris out much faster. But as it comes out, it blocks me from easy access so I wipe it off.

    I recleave because my fiber gets damaged on its distal end, and I want max power coming out so I can move the procedure along. I don’t recleave due any concern of a hot glass-like shift. But I love the reasoning!!smile.gif

    Bob

    #8183 Reply

    Robert Gregg DDS
    Spectator
    QUOTE
    Quote: from Swpmn on 7:13 pm on June 30, 2004

    QUOTE
    At 650 one needs to move pretty fast or have a massive bleeder.

    Bob

    I’m confused by this statement and need clarification.

    Do you mean?:

    1) In most instances, if the practitioner is using the Periolase at 650 usec PD he/she needs to move the fiber tip rapidly to avoid tissue damage(because the laser is firing for a long time at 650 usec)

    2) However, the 650 usec setting can also be used in a focused state when the practitioner has a large blood vessel with difficult to control hemorrhage

    Thanks,

    Al

    Yeah, that’s about it………..Things happen quickly at 650, both good and bad if one isn’t careful and trained.

    Or the operator can turn the PD down to say 350 and not have to move the tip quite as fast.

    Bob

Viewing 15 posts - 31 through 45 (of 64 total)